so complex that their use has been confined to research
settings. Recently, washout/washin techniques using
oxygen [6,7] or nitrogen [8,9] have been made available
in ICU ventilators, allowing bedside EELV measurement.
A comparison of the nitrogen washout/washin EELV
measurement with helium dilution or CT scan had
shown good correlations in stable patients ventilated
with low-PEEP levels [8]. The limitations of the nitrogen
washout/washin technique for EELV measurement
under other conditions, such as high FiO
2
or high PEEP,
have not been fully investigated [10].
PEEP-induced changes in lung volume (referred to as
PEEP-volume) can also be assessed simply at the bed-
side by using passive spirometry. This accurate method
requires a long expiration to zero end-expiratory pres-
sure (ZEEP), where FRC is assumed to be reached.
When considering the changes induced by two different
levels of PEEP in a given patient, the difference in EELV
(that is, ΔEELV = EELV
high PEEP
-EELV
low PEEP
) should
theoretically be similar to the difference in ΔPEEP-
volume (PEEP-volume
high PEEP
– PEEP-volume
low PEEP
),
assuming that the FRC has not been modified by the
PEEP changes (see Figure 1).
We designed a multicenter study with the primary
objective of investigating the precision (reproducibility)
of the nitrogen washout/washin technique for EELV
measurement in patients with ALI/ARDS at two PEEP
levels, including a high level, with a small variation in
oxygen concentration (10%). Our secondary objective
was to evaluate the accuracy of the technique by
comparing PEEP-induced changes (Δ)inlungvolume
with the nitrogen technique or the PEEP-volume above
the FRC measured with passive spirometry. As PEEP-
volume is relatively easy to measure accurately with a
calibrated pneumotachograph, it may therefore be con-
sidered a “gold standard.”Because we expected possible
discrepancies between the two techniques, we also com-
pared the measured changes in lung volume (ΔEELV
and ΔPEEP-volume) with the minimal predicted
increase in lung volume, computed from static compli-
ance (C
stat
) at low PEEP and ΔPEEP. The minimal pre-
dicted increase in lung volume was considered the
smallest-volume increase that can occur. We have also
used this method to evaluate alveolar recruitment, as
described elsewhere [11].
Material and methods
This was a multicenter study performed in five French
medical intensive care units at the Henri Mondor Uni-
versity Hospital in Créteil, European Georges Pompidou
University Hospital in Paris, Angers University Hospital
in Angers, l’Archet 1 University Hospital in Nice, and
Charles Nicolle University Hospital in Rouen. In compli-
ance with French legislation, the institutional review
board of the Henri Mondor University Hospital
approved the protocol for all centers and waived the
need for informed consent, as PEEP optimization was
considered part of standard care. The patients or next
of kin received information about the study.
Dellamonica et al.Critical Care 2011, 15:R294
http://ccforum.com/content/15/6/R294
Page 2 of 8